Sample itemA patient with an upper arm circumference of 40 cm has a non-invasive blood pressure of 172/104 mmHg on a standard adult cuff, while an arterial line shows 138/82 mmHg. What is the most likely explanation?
Original practice item, not a recalled exam question. Nothing here is recorded.
The answer and the reasoning
D. Cuff too narrow for the arm
A cuff that is too narrow does not transmit pressure evenly to the artery, so the reading is falsely high. The cuff width should be about 40% of the arm circumference, with the bladder encircling at least 80% of the arm. Too wide a cuff, rapid deflation and a high arm all give falsely low values.
Option by option
A. Cuff placed above the level of the heart
A limb above heart level reads lower because of hydrostatic pressure.
B. Cuff deflated too quickly
Rapid deflation tends to give inaccurate and low systolic readings.
C. Cuff too wide for the arm
A cuff that is too wide underestimates blood pressure.
D. Cuff too narrow for the arm (correct)
Correct. A narrow cuff overestimates blood pressure.
Source note: Original NextPrep sample item — pending expert review. This item is original practice material awaiting independent expert review, so confirm important facts in a standard textbook.
More Anaesthesia questions
- Monitoring and anaesthesia workstation: curriculum recognitionA catheter is placed through the internal jugular vein with its tip in the lower superio…
- Monitoring and anaesthesia workstation: curriculum recognitionOn a five-lead ECG in theatre, which lead is the most useful single lead for identifying…
- Monitoring and anaesthesia workstation: curriculum recognitionAn anaesthetist wants to stop a vaporiser calibrated for one volatile agent from being t…
- Monitoring and anaesthesia workstation: curriculum recognitionDuring volume-controlled ventilation of an anaesthetised patient, the low expired minute…