Radiology · Imaging modalities and physics: off-level grid cutoff
Sample itemA radiograph taken with a focused grid shows an even loss of density across the entire film. The tube was angled slightly across the grid lines, and the focus-grid distance was correct. Which fault is this?
Original practice item, not a recalled exam question. Nothing here is recorded.
The answer and the reasoning
C. Off-level grid cutoff
When the central ray is not perpendicular to the grid plane (a tilted tube or tilted grid), the lead strips absorb primary photons across the whole field, giving uniform density loss. Off-focus cutoff, in contrast, darkens the centre and loses density towards the periphery.
Option by option
A. Moire pattern
Moire patterns arise when a stationary grid interacts with the scanning frequency of a computed radiography reader.
B. Grid line artefact from a stationary grid
A stationary grid leaves fine parallel lines on the image and does not cause an even loss of density.
C. Off-level grid cutoff (correct)
Correct. A tilt of the grid relative to the beam causes uniform loss of density across the film.
D. Off-focus grid cutoff
Off-focus cutoff arises from the wrong focus-grid distance and loses density mainly towards the edges.
Source note: Original NextPrep sample item — pending expert review. This item is original practice material awaiting independent expert review, so confirm important facts in a standard textbook.
More Radiology questions
- NeuroimagingWhat is the first investigation of choice in suspected acute subarachnoid haemorrhage?
- Cardiac imagingA 'boot-shaped' heart (coeur en sabot) on a chest radiograph is typical of:
- Musculoskeletal imagingA 'bamboo spine' on a plain radiograph is characteristic of:
- Paediatric radiologyThe 'double bubble' sign on an abdominal radiograph of a newborn suggests: